- Wall-mounted oxygen sensor for use in combination with measurement monitor (for 1 point).
- This oxygen sensor uses a limit-current type zirconia oxygen sensing element, and is powered by DC voltage supplied from outside or in combination with a dedicated monitor MG3010.
- The MG2100 can also perform measurement display, alarm and transmission signal output operations.
- zirconia oximeter
- Stabilized zirconia is known as an oxygen ion conductor.
- The stabilized zirconia is used as a substrate, and a platinum electrode and an oxygen diffusion limiting hole are formed on one surface of the substrate, and a platinum heater is formed on the other surface.
- In order to obtain practical oxygen ion conductivity, the substrate is heated to several hundred °C by energizing the heater. When a voltage is applied between the electrodes, a current with oxygen ions as carriers flows through the substrate.
- Due to the oxygen diffusion limiting hole formed on the cathode, the amount of oxygen flowing into the cathode is limited, and a flat region (limit current) is observed in the current-voltage characteristics.
- Since the limiting current varies in proportion to the oxygen concentration, it is possible to determine the oxygen concentration in the atmosphere by measuring the limiting current.
ข้อมูลจําเพาะ
- Transmission signal output: 0~10 mV (output resistance approx. 10 Ω)
- alarm setting
- Densitometer Mode Type II (fixed by user)
- Lower limit value, Upper limit value, buzzer enabled/disabled, lamp enabled/disabled, relay output enabled/disabled, output holding enabled/disabled
- Unspecified parameters are the default values for densitometer mode Type I.
- Display range: 0.00~2.50%O2 fixed
- Transmission output range: output corresponding to 0.00~2.50%O2
- *Possible combinations of display range and transmission output range are 00, 11, 20, and 22. However, only 00 can be specified when the alarm setting is the oxygen deficiency meter mode.
- For densitometer mode Type I and display range 1: Alarm initial value is lower limit 0.00%, upper limit 2.50%
- Measurement Principle: Zirconia Solid Electrolyte Method
- Air sampling method: natural diffusion type
- Response time: 90% response within 20 seconds
- Oxygen detection element position: fixed to the dedicated holder in the main body case
- Measurement value display: Digital display with 3-digit yellow-green LED
- Measuring range (resolution): 0.00~2.50 (0.01)%O2
- Accuracy rating: +/- 0.05%O2 +/- 1digit in a dry gas of oxygen and nitrogen
- Transmission method: output to 0.00~2.50 measurement indications
- Alarm function: Measured value alarm ...
- Alarm setting code: Densitometer mode *1
- *1 Alarm setting "A" is set to the initial value at the time of shipment, and alarm setting "B" is set to the state of alarm setting "0" for unspecified alarm value and alarm output operation.
- Lower warning value: specified within 0~26% *2
- Upper warning value: specified within 0~26% *2
- *2 The lower alarm value cannot be set to a value larger than the upper alarm value or cannot be specified. When alarm release alarm is held, it is released by front reset key.
- Relay output yes/no: Optional
- Buzzer with/without: Optional
- Lamp ON/OFF: Optional setting
- Output Hold Yes/No: Optional
- Detection element abnormality alarm: When the heater current to the oxygen detection element is abnormal, relay output is ON, buzzer is blown, and lamp is lit regardless of the content of the measured value alarm.
- Power failure warning: When the power supply to the tester is OFF or when the fuse is broken, relay output is ON.
- Alarm relay output: number of output circuits (2), output capacity (Resistive Load 110 V AC 0.2 A, 24 V DC 1 A/Inductive Load 110 V AC 0.1 A, 24 V DC 0.5 A)
- Warm-up time: 5 minutes
- Note)Alarm signal is OFF during warm-up, transmission signal outputs minimum value
- Operating Temperature Range: 0~40°C
- Operating humidity range: 80%rh or less
- Power: 4.2~5.5 V DC
- When power is supplied from the MG3000 series, use a 2mm2 lead wire within 30 m of each side. Use a 0.75 mm2 lead wire and connect it to a length of 11 m or less per side (wire resistance of 0.3 Ω or less per side). If this length is exceeded, a 100 V AC power supply is required, but use MG1000.
- Power Consumption: Approx. 3VA
- Weight: about 450 g
- Mounting method: wall or switchbox mounting
- ISO 14001 Certified JQA-EM2414
- ISO 9001 Certification JQA-0656
ขนาดแพ็คเกจ:120×55×120 mm 450 g [เกี่ยวกับขนาดแพ็คเกจ]
| หมายเลขใบสั่ง | 67-4915-53 | |
|---|---|---|
| หมายเลขแบบจําลอง | MG2101-B11 | |
| รหัส JAN | 4562208231397 | |
| ราคามาตรฐาน |
JPY: 173,000
USD: 1,084.44
Excange rate 1USD= 159.53JPY
Valid price in Japan |
|
| ปริมาณ | 1piece | |
| หุ้นในประเทศญี่ปุ่น |
|
|
| หุ้นซัพพลายเออร์ |
|
|
ผลิตภัณฑ์ที่เกี่ยวข้อง (ขนาดที่แตกต่างกัน, ข้อมูลจําเพาะ,ผลิตภัณฑ์ตัวเลือก ฯลฯ)
| รูปภาพผลิตภัณฑ์ | หมายเลขใบสั่ง | ชื่อ | หมายเลขแบบจําลอง | ปริมาณ | ราคามาตรฐาน | ราคาที่ถูกต้องในประเทศญี่ปุ่น | หุ้น [หุ้นซัพพลายเออร์] |
|
|---|---|---|---|---|---|---|---|---|
![]() |
67-4915-46 | Mg2100Series Wall-Mounted Oxygen Sensor (Transmission Signal: 0~10 Mv, Acidity Meter Mode I, Display: 0.0~25.0%O2 Fixed, Transmission Output: 0.0~25.0%O2 Supported) MG2101-000 | MG2101-000 | 1piece | JPY: 149,000 | USD: 933.99 |
|
|
![]() |
67-4915-47 | Mg2100Series Wall-Mounted Oxygen Sensor (Transmission Signal: 0~10 Mv, Acidity Meter Mode Ii, Display: 0.0~25.0%O2 Fixed, Transmission Output: 0.0~25.0%O2 Supported) MG2101-100 | MG2101-100 | 1piece | JPY: 152,000 | USD: 952.80 |
|
|
![]() |
67-4915-48 | Mg2100Series Wall-Mounted Oxygen Sensor (Transmission Signal: 0~10 Mv, Densitometer Mode I, Display: 0.0~25.0%O2 Fixed, Transmission Output: Supports 0.0~25.0%O2) MG2101-A00 | MG2101-A00 | 1piece | JPY: 152,000 | USD: 952.80 |
|
|
![]() |
67-4915-49 | Mg2100Series Wall-Mounted Oxygen Sensor (Transmission Signal: 0~10 Mv, Densitometer Mode I, Display: 0.00~2.50%O2 Fixed, Transmission Output: Supports 0.00~2.50%O2) MG2101-A11 | MG2101-A11 | 1piece | JPY: 168,000 | USD: 1,053.09 |
|
|
![]() |
67-4915-50 | Mg2100Series Wall-Mounted Oxygen Sensor (Transmission Signal: 0~10 Mv, Densitometer Mode I, Display: 0.0~25.0%O2/0.00~2.50%O2 Automatic Switching, Transmission Output: Supports 0.0~25.0%O2) MG2101-A20 | MG2101-A20 | 1piece | JPY: 165,000 | USD: 1,034.29 |
|
|
![]() |
67-4915-51 | Mg2100Series Wall-Mounted Oxygen Sensor (Supports Display Of Transmission Signal: 0~10 Mv, Densitometer Mode I, Display: 0.0~25.0%O2/0.00~2.50%O2 Automatic Switching, Transmission Output: 000~250) MG2101-A22 | MG2101-A22 | 1piece | JPY: 168,000 | USD: 1,053.09 |
|
|
![]() |
67-4915-52 | Mg2100Series Wall-Mounted Oxygen Sensor (Transmission Signal: 0~10 Mv, Densitometer Mode Ii, Display: 0.0~25.0%O2 Fixed, Transmission Output: Supports 0.0~25.0%O2) MG2101-B00 | MG2101-B00 | 1piece | JPY: 157,000 | USD: 984.14 |
|
|
![]() |
67-4915-53 | Mg2100Series Wall-Mounted Oxygen Sensor (Transmission Signal: 0~10 Mv, Densitometer Mode Ii, Display: 0.00~2.50%O2 Fixed, Transmission Output: Supports 0.00~2.50%O2) MG2101-B11 | MG2101-B11 | 1piece | JPY: 173,000 | USD: 1,084.44 |
|
|
![]() |
67-4915-54 | Mg2100Series Wall-Mounted Oxygen Sensor (Transmission Signal: 0~10 Mv, Densitometer Mode Ii, Display: 0.0~25.0%O2/0.00~2.50%O2 Automatic Switching, Transmission Output: Supports 0.0~25.0%O2) MG2101-B20 | MG2101-B20 | 1piece | JPY: 170,000 | USD: 1,065.63 |
|
|
![]() |
67-4915-55 | Mg2100Series Wall-Mounted Oxygen Sensor (Supports Display Of Transmission Signal: 0~10 Mv, Densitometer Mode Ii, Display: 0.0~25.0%O2/0.00~2.50%O2 Automatic Switching, Transmission Output: 000~250) MG2101-B22 | MG2101-B22 | 1piece | JPY: 173,000 | USD: 1,084.44 |
|